Have anyone noticed a drift in an LC-MS-MS method despite the use of an isotope as internal standard (IS)? I run a method with several analytes - all with a corresponding isotope as internal standard. When I recently tried to run the method on another (equivalent) LC-MS-MS system, I noticed a drift of two of the analytes. The concentration increased in both cases, but in the first case it was due to decreasing IS area (the analyte area stayed the same) and in the other case it was due to increasing analyte area (the IS area stayed the same). Can anyone explain this? Several of the other analytes are OK. The only (obvious) thing that unite these two analytes is that they both are the first channels in their respective MRM-function (I use two MRM-functions). Is ion suppression or ion enhancement possible, despite a co-eluting internal standard? The ion source was cleaned the same day - may a unoptimal vacuum give rise to these problems?