Page 1 of 1

How to select "loop final pressure" on Agilent 7697A

Posted: Mon Aug 09, 2021 1:49 pm
by bry_fieri
How to select "loop final pressure" on Agilent 7697A headspace? Should the final loop pressure be higher or lower than the carrier pressure?

Re: How to select "loop final pressure" on Agilent 7697A

Posted: Fri Aug 13, 2021 2:37 pm
by DANIELECRD
Final pressure is not related to carrier pressure. You should choose a pressure lower than that used to pressurize vial: this difference is used by system to fill loop with sample. initial pressure is dependent from sample matrix and method condition: for example you can start from 15 psi and finish at 11.
Try to find best combination for your samples.