Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

11 posts Page 1 of 1
Evaluating the use of O-phosphoric acid in a MP phase prep. The MP is prep'd @ 60 mM phosphoric then pH adjusted to 4.4 w/ NaOH. Method is isocratic, 1 mL/min, 220 nm w/ 15 min stop. Sx's prp'd in 0.05% sodium metabisulfite.

Problem: The baseline has a very consistent "hump" that starts @ 2 min (shortly after void) and ends @ 12 min w/ the baseline from 0-2 minutes and the baseline from 13-15 min being very stable w/ ~ 0 absorbance.

Any thoughts?

tnx
shawn

Aside from the fact that phosphate isn't a buffer at 4.4 ?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

Tom-
Am trying to replace pyrophosphoric acid (pK 5.77) which shouldn't really buffer @ 4.4 either. However, the method is very rugged and provides excellent chrom.. My leader would like for me to stick w/ an inorganic acid for buffering, and none of them have a pK @ 4.4 (HF and tetraboric are closest, but would like to not use them if possible). So I was just evaluating the phosphoric for "fun", and was suprised to see the resultant chrom. Even if the solution is not properly buffered, what would be causing the "hump"?

Shawn,

Do you see the hump only when you are injecting your samples or even with blank runs (injection of water). Does your hump change with different injection volumes?

mtnshawn-

my speculation/concern was that without effective pH control, something in your sample or matrix is smearing out to give the baseline problem.

Kostas' suggestion should shed some light on that question (if you see the same problem with a blank injection)
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

Blanks (H2O) have humps. Diluent (0.05% sodium metabisulfite) has humps. Humps are also seen on brand new column.

One potential clue, is that during equilibration the baseline is great, so you would think that it is injection related. However, when I perform "fake" injections I don't get the humps. Leading me to believe that the injection pathway (loop, rotor seal, lines leading to the column) is clean.

The needle has been thoroughly cleaned (50 injections @ 50 uL) w/ 0.5% phosphoric in an attempt to get what ever contaminants there may be out of the needle.

Talk about having fun.
sc

From the sparse information, and abreviations I couldn't figure out, it was difficult to think too deeply about this problem, but one comment caught my eye.

What about typical chromatography challenges a buffer pH to any great extent? Unless the sample pH( and we learned nothing about the sample from this question. While, all the available evidence points to the sample as the problem) is buffered greatly different from the mobile phase, it should take but a smidgen of buffer capacity to control pH during elution. A buffer prepared 2, or even 3 pH units from its pKa provides at least a smidgen of buffer capacity. So the original buffer (pKa 5.8 and buffer pH of 4.4) provides lots of buffering capacity unless the sample pH offers a challenge to this capacity.

Why is bisulfite present. Is it a chromophore at 220 nm. Were does it elute?
Bill Tindall

Bill-
bisulfite is present because this is a modified* USP method for an API (Active pharmaceutical ingredient) and the USP uses the bisulfite to minimize oxidation of the API. We have chosen to keep the metabisulfite in the preparation in order to not deviate to much from the USP method.

Modified*-due to our formulary matrix, we were unable to run the exact USP method and obtain adequate resolution between the API and the preservative. By dropping the pH to 4.4, resolution issues were resolved.

It does have some chromophoric capacity @ 214 nm. It elutes @ the void.

At this point I haven't even evaluated formulary samples. All the work has been done on an analytical standard (purity > 98%) prepared in the 0.05% metabisulfite.

Not sure which abbreviations were causing problems, but here is a list.
MP - mobile phase
sx - sample
sc

I'm stumped! Any hypothesis I had was ruled out by experiments you have already done, especially the water blank injection. It seems like water is displacing something from the coumn, but what? And the problem does not occur with pyrophosphate? Let us know if you figure it out.
Bill Tindall

What's the column? RP? if RP, which one? Is there anything else in the mobile phase, or just your salt and water?

Have you tried doing an injection of the mobile phase (instead of water or sample solvent)? If so, does the hump still appear?
11 posts Page 1 of 1

Who is online

In total there are 26 users online :: 1 registered, 0 hidden and 25 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 25 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry