I should make the case more clearly. The active pharmaceutical ingredient API(A) and API (B) are two different compounds. API(A) is sodium salt and unstable at pH <3. API (B) is sulfate salt and unstable at pH>8.
I tried to separation API(A) and API(B) at pH=9 with a YMC Pro-Pack, C18 column, but no separation was achieved. Since I am also concern with the degradation of compound B at high pH, I chose to use acidic mobile phase to satisfy the separation first, and meanwhile to overcome the on-column degradation of compound A.
Initial condition: C18 column 150x 4.6 mm,3u,isocratic, mobile phase of acetonitrile: water: TFA (70:30:0.05), pH =1- 2, flow rate at 1ml/min, API(B) elutes faster than API(A). Under this condition, I can separate A from B and other 7 known impurities which reached my first objective. However, an on-column degradation impurity peak from API(A) was detected.
To confirm the impurity peak is an on-column degradation product, I tested API(A) alone under another chromatographic condition (pH=5-6, compound A is stable, and it is well separated from the known degradation impurity) with a shorter column, then compared the chromatogram obtained from previously described conditions. No degradation impurity peak was observed as using the less acidic mobile phase. The same sample was injected onto the pH=2 mobile phase. I found the baseline is elevated after the elution of compound A, and the degradation impurity peak (<0.4%) was eluted. It appeared that the degradation happened as soon as API(A) was exposed to acid, the API (A) and the degradation product move along the column and are separated.
As Bill suggested, the key point to control the on-column degradation is time and temperature. All tests were performed at room temperature. I do not have column chiller so lower temperature was not further pursued. I will try a short column, such as 10 cm or 7.5 cm column to reduce the run time. Any suggestions? Thanks.