I have been struggling for a couple months, on and off, to get our new GC up and working to analyze for USP <467>.
The column, temp program, and gas pressures are as specified in the USP monograph. Actually I'm doing everything they say, with the exception of Headspace operating parameters. Instead of 30sec pressurization time, I'm using 2 minutes because that was recommended by the HS40 user manual.
The problem is, for the Class 1 mix, using Procedure A (haven't even gotten to "B" yet...) I can't see any kind of peaks eluting. I've tried turning the detector attenuation to -5, but that only makes it a really amplified, noisy baseline. For Class 2A, the peaks come out nicely.
Does anyone have some suggestions for fine-tuning? Anybody working with PE equipment (I've got a Clarus 500 and a HS40 headspace unit. Both are new so unlikely to be malfunctioning)
I'd appreciate any and all help or ideas!