I developped a method for 2 basic compounds using LC-ESI-MS/MS. Separation is done on a C18 column using a high pH mobile phase (MeOH/ H2O 25 mM ammoniumbicarbonate pH 9.5).
I analysed urine samples (which are postmortems, so quite dirty) with mixed mode SPE (Bond Elut Plexa PCX). Recovery is good (70-90%), matrix effect is fine for one component (85%, where 100% is no matrix effect), but not good for the first-eluting compound (16-40%).
I tried following things, but they didn't improve the results.
* low pH mobile phase
* different gradients with the high pH mobile phase in order to separate the analyte from interferences, this didn't really help, exept when using a 30 (!) minute run...
* I lowered the flow rate from 0.5 ml/min to 0.25 ml/min.
I was thinking to change the column chemistry (maybe try a phenyl column) or changing column length in order to get more resolution...
Since I still have to purchase these columns, I was wondering if you could give me some advice about what could be the best option... Or are there other ways to avoid the matrix effect?