I have a problem with Varian saturn2000/3800 equipped with 1079 during pah analysis.
All heaviest PAH doesn't exist in my chrom!
Until PAH with mass 202 the sensitivity is ok, but PAH from m/z 252 to 302 completely disappear.
I have also cleaned the trap, changed column (rxi-5sil 30m*0.25*0.25) and liner (1079 fritted liner) but no change happens.
The only thing that give some increment to signal was temp rising of trap from 220°C to 230°C
Some other condition are
InJection: large volume of 9uL, solvent dicloromethane
flow 1.3mL of helium
trap temp 230°C
t.line temp 300°C
Any advice are greatly appreciated..
Thanks to all
[/img]
daniele
------------------