I'm not sure what to point a finger at here, as before I ran a calibration on this instrument, it passed the MWD self-test, Cell Test, Filter Test, Pressure Test, Leak Test, column compartment temperature test, Wavelength calibration test, etc. This instrument, as far as I could tell, was tip-top. This morning I switched water sources (was running store-bought purified water, switched to water from our recently serviced Millipore unit, still have the baseline issue). Degassing shouldn't be an issue, as the degasser is a recently refurbished (by Agilent) unit. Even so, I switched the channel in which I have my water being degassed, to know avail. I also flushed the flow cell for 5 minutes at 4 mL/min with water and then 52:48 water:methanol in case it was an air bubble or other contamination. This instrument is a recent acquisition, so I don't know its particular history, but I'm running two other instruments with the same mobile phase, and one with the same type of column but a DAD detector, and I don't have this issue.