different peak size?

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
I have a vapor sample that has the sme conc of benzene and m-xylene, but the area counts for these two chemicals are quite different? Could anyone tell me why this would happen?
Thanks!

Hong

What type of injection (split, splitless) are you using and your method of detection? While there are many possible answers to your question, inlet discrimination and differences in detector sensitivity are common problems. With a mass spectrometer, it makes a difference as to wether you take the signal for a single ion (and which one) or the TIC. A description of your instrument configuration and operating conditions can help.