In an analysis of naphthalene in honey I observe a peak with the retention time of naphthalene. This peak has a considerable area. The area is about a half of the LOQ.
Initially I thought that it is a classic case of carryover, but subsequently found that it is not so.
I tried to change: acetonitrile with methanol, a type of C18 column, the ratio of acetonitrile in mobile phase. Peak is still there. Using the gradient also failed to alter the presence of this peak.
I noticed that the purity of the mobile phase is crucial. Moreover, attempts to use a plastic ware for sample pretreatment greatly worsen the situation. I used plastic ware from various suppliers. Peak becomes even higher.
Finally, another instrument was used in another laboratory. Of course, I changed everything: supplier of solvents, glassware ... everything.
And guess .... YES ... the peak was there.
Currently, I have to work with glassware and wash it in advance with the mobile phase, which complicates the analysis.
I am wondering if anyone has observed a similar phenomenon.
The HPLC conditions are:
Mobil Phase: AcCN:H2O (50:50, v/v);
Column: C18 4.6 mm x 150 mm, 5 um;
Flow rate: 1.5 mL/min;
Detection: Fluorescence Excitation 275 nm, Emission 330 nm;
Injection volume: 50 uL;
Sample and standards are dissolved in mobile phase.
Thanks in advance for your opinions and findings.