How I clean a flow cell?
Posted: Wed Aug 26, 2009 12:13 am
by koko
Hello everyone!
I have problems with the detector Waters 2487, the signal is unstable, I cleaned it backflushing water at 0.2ml/min all night but the signal still unstable, the recommendation of Waters is that remove and cleaned it with Nitrogen but I have no nitrogen in the lab. What can you recommend?

Posted: Wed Aug 26, 2009 12:52 am
by tedinelkgrove
you could flush it with a dilute nitric acid solution, or clean the lenses with IPA and kim wipes.
Posted: Wed Aug 26, 2009 8:25 am
by danko
Hi Koko,
In my terminology “unstableâ€
Posted: Wed Aug 26, 2009 11:31 am
by grzesiek
have you seen it? I mean the flow cell?
Posted: Wed Aug 26, 2009 2:54 pm
by koko
danko:
Yea, when I said unstable I tried to meant varying, so, what you recommend?
grzesiek:
No, I have seen the flow cell!
tedinelkgrove:
Do you mean that I remove the flow cell and only clean the exterior?
Posted: Wed Aug 26, 2009 3:15 pm
by grzesiek
"No, I have seen the flow cell! " - you mean you have not seen it?
the problem as recognized by danko doesn't have to be dirty flow cell, can be for example air in the detector, mixing problems etc
"I cleaned it backflushing water at 0.2ml/min all night" - purging strong solvent through could be a better option
Posted: Wed Aug 26, 2009 3:26 pm
by danko
Hi Koko,
In order for me/us to suggest potential solutions, it is important for you to describe the problem in more unequivocal manner.
F. ex.:
1. How does the signal instability manifest its self?
2. Do you see fluctuating, edgy baseline?
3. Or do you experience varying peak area (i.e. height, width or both) when running your analyte?
In any case, until further notice, you don’t need to attempt any flow-cell cleaning procedures. Because dirty cell would result in lower signal and potentially more pronounced noise. But it’s quite unlikely to cause variations as such.
Best Regards
Posted: Thu Aug 27, 2009 9:29 pm
by koko
I will use THF as a strong solvent to flush the cell. Before I flush the flow cell I had no response of the sample, after flush it I have response but the baseline is fluctuating.

Today I flush it with HNO3 6N and then water like the manual said. And the baseline still fluctuating.

Posted: Fri Aug 28, 2009 6:40 am
by danko
Hi again,
As mentioned earlier, I don’t think flushing the flow-cell will bring you any closer to the desired situation i.e. stable baseline. Actually if you have an air bubble in the flow-cell (it’s easy to determine, by the look of the baseline) a reasonably hydrophobic liquid, such as ACN, or THF for that matter, would easily remove it. So, unless you’re getting found of flushing for the fun of it, you may consider some other explanations of the problem you’re dealing with.
One such might be flow/pumping problems. But if you make the effort to post a couple of data examples (a chromatogram and a baseline alone) it may help a lot.
Best Regards
Posted: Fri Aug 28, 2009 8:41 am
by HW Mueller
Weird . . . ., don´t the detectors have selftests anymore? Unstable with flow and/or no flow? Anything systematic possible here? What do you mean with unstable anyway? What is the nitrogen you don´t have?
Posted: Mon Aug 31, 2009 9:29 pm
by vofelyp
Hi Koko
What do you mean unstable? Is it cyclic or not? What is the cycle time? Short cycle time noise come from the pump, long cyclic noise come from other source most frequently from the aircondition device or the online degasser. Non cyclic noise dirty eluent, not good gradient mixing.
You can check the lamp hours and number of ignition (Shift + 1). After 2000 hours the lamp energy decrease quickly.
The lamp energy check:
Fill the cell with MeOH, and set 230 nm. In the diagnostics menu choose the Sample & ref energy menu. The reference energy should bigger 40, if not the lamp and/or optics has failed.
Posted: Thu Sep 03, 2009 11:31 pm
by koko
Thank you everyone, I found the problem, the lamp had 2800 hours.