The configuration is a two column heart-cut with backflush.
The nitrogen elutes at about 1/4th the time of krypton, about 12 seconds at the beginning of the heart-cut, the cut being about 6 to 12 seconds in width, and then the backflush is operated.
You may be able to do a simple backflush to vent, but I suggest that you don't for best results. If it doesn't give you a high background with your detector, go ahead and do a simple backflush at about 30-35 seconds.
Other details you can discuss with Supelco Technical Service.