In the past we didn't have any problems with DDT-breakdown, but since a few weeks we see that during a series DDT in the first standards is OK, but 15 or 20 injections later no peak is found anymore.
We use the same type of liners and analytical column since we started. The guard column was replaced by another type, but the new type is also used by another pesticide analysis we do without any problems (that analysis uses cold on-colum injection) and our suplier says that this type should be at least equal to the type we used before.
The frequency of replacing liners, guard columns and analytical columns hasn't changed (or even increased since we have this problem).
We do not use press-fits, but use a self-made system in which the analytical column is put into the guard column for about 1 cm. This connection is used for years and didn't give any problems so far and is still working on other applications in which pesticides are measured.
The problem occurs on both GC's on both channels.
Any ideas on the things I could check now?