i've tried acetate buffer.. phosphate buffer.. both with and without salt.. with and without 10% ethanol.. my protein gives a lot of tailing.. this is on a sillica matrix tsk columns.. any suggestions??
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
yess.. how many injections does yr sec column survive generally?The use of both columns is rather low for explaining problems.
How do you measure recovery?
i use some 7.8*78 dimension column.. dont remember exactly but near this.. and what i do is make a 0.3 mg/ml solution from around 0.6 mg/ml in mobile phase.. and inject 100 microlitres of it.. u recommend injecting lower volumes? isnt dilution in mobile phase prefered for better recovery? do i need higher concentrations of the starting solution? i can get that.. wat say!? thanksss!!This could primarily be a matter of integration. The peaks are relatively broad in SEC, they disappear relatively fast in noise as you lower the injection amount.
Also, the volume you inject should be around 5µl.
How is the response when you inject without a column (same flow rate)?
Furthermore, you may have to "live" with some tailing, which, of course, contributes to bad integration at low levels.
i dilute 50-fold yes.. so i wud try lower injection volumes.. and i have gone from 100-200 mM salt.. i wud try higher.. i have tried higher pH 5.2 to 7.2 i cant go higher there.. and i have tried 10%ethanol.. the tailing increases.. indicating hydrophobic interaction not being a cause. also.. i have been trying acetate and phosphate buffers for these conditions.. wat else?You make a 100 microL injection for the 30 microg injection. Do you make a 2 microL injection for the determination of the peak area for the 2% value?
If you dilute the sample 50 fold, a possible source of loss is the sample vial. Proteins love to stick to glass...
It is not clear if the tailing is from silanols or from other features of the packing. For silanols, an increase in the concentration of the salt would help. If it is due to other interactions, you need to sort them out one by one. One possibility is hydrophobic interaction, which does not go away with the addition of salt, but would get worse. If this is a possibility, you can add some 5% or 10% methanol. Don't know what this will do to your assay of the aggregates, but it will help sorting out the cause of the tailing.