how short is too short for a guard column?
Posted: Thu Aug 13, 2009 7:39 pm
were running astm d6584 method for analysis of glycerides in biodiesel.
the method recommends a 2-5m guard to extend column life. weve been starting with a 2m guard and clipping it down to say, 1/2 meter before discarding
triglycerides are notorious for gunking up a column, and were using an expensive hi-temp metal column. I dont want to push it and risk column damage by using too short of a guard.
any recommendations? Is there any disadvantage to starting with a longer 5m guard? i imagine since its deactivated the added heat cycling wont really affect anything, all the contamination tends to occur in the first few inches anyway, which gets progressively clipped
thanks!
the method recommends a 2-5m guard to extend column life. weve been starting with a 2m guard and clipping it down to say, 1/2 meter before discarding
triglycerides are notorious for gunking up a column, and were using an expensive hi-temp metal column. I dont want to push it and risk column damage by using too short of a guard.
any recommendations? Is there any disadvantage to starting with a longer 5m guard? i imagine since its deactivated the added heat cycling wont really affect anything, all the contamination tends to occur in the first few inches anyway, which gets progressively clipped
thanks!