Dear all,

Is there anyone touch the LC analysis of Vinorelbine tartrate? I read the USP monograph of vinorelbine tartrate for our HPLC control method and the context indicates the column(i.d.3.9x150mm)packing is 5 micrometer L1 in the assay and related compound sections, however,as many chromatographers realize that different column manufactures make the variation of different brand columns, which leads impact on chromatography performance, and I found some literatures showed that Waters Nova Pak, Agilent Luna C18 columns are used for this comound, but I'm not sure which is the best, so I hope you give me helpful information of the suitable brand packings(the brand of the submitter is preferred). Thanks in advance! :o

Austin
[/code]
The God had ever have three apples. Adam was tricked eating up one of them in Eden, and the second dropped from the tree and hit Issac Newton on his head, then what happened on the third one?Interestingly it was bit by Steve Jobs!

It depends on what the HPLC analysis really is! If it’s some kind of purity test (with difficult separations) then you’ll need to find out which brand is most suitable. But if it’s “onlyâ€
Learn Innovate and Share

Dancho Dikov

With some of these things, you can read between the lines and get the information that you want. A 3.9 mm i.d. column is ALWAYS a Waters column, and in this case I think that Novapak is the correct one, since uBondapak does not come in a 5 micron version.

that's a common problem and solution is simple - contact person responsible for the monograph and ask :)

Thanks everyone!

I've checked WATERS updated column catalog and it lists Novapak i.d 3.9x150mm column is filled with 4 micrometer packing only, and 5 micrometer particle size columns such as XTerra, Atlantis, Symmetry brands are available. And WATERS' local engineer told that particle size make little impact on separation with constant other column conditions and I will develop my method with the XTerra column in hand and study the chromatograms, and then discuss the further actions.

And I commuicated with the person who is responsible for the monograph and he will give me the feedback one week later. I hope he will give me a hand greatly. :lol:

Austin
The God had ever have three apples. Adam was tricked eating up one of them in Eden, and the second dropped from the tree and hit Issac Newton on his head, then what happened on the third one?Interestingly it was bit by Steve Jobs!
Dear Austin,
in the Ph. Eur. your product is written with a LC-colum Deltapak, 15 * 3,9 mm, 5 µm, 7 % carbon load (Waters Corporation).
I think, the USP should be similar or equal.
Regards!
Antje

Antje, thank you very much!:) :P

As your answer, the USP's correspondence told me that the method was validated with the Delta-Pak HPI C18 column, 5 micron, 3.9x150mm.

Thanks everyone for your help! :alien:
The God had ever have three apples. Adam was tricked eating up one of them in Eden, and the second dropped from the tree and hit Issac Newton on his head, then what happened on the third one?Interestingly it was bit by Steve Jobs!
7 posts Page 1 of 1

Who is online

In total there are 15 users online :: 0 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 15 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry