Short story: all gases and compounds even up to benzene are non-detect at any concentration level. Peak shape gradually improves from heavy tailing to very sharp throughout the run.
Longer story: this lab has the instrument set up a little differently than what I am accustomed. Using double tapered splitless liner and running totally splitless, their chromatography has been ok for gases until recently.
I removed the liner and it was packed to the gills with septa shavings and replaced with another double tapered splitless. I was hoping this would solve the problem but it is worse now by a lot.
Direct injection produces good chromatography if I turn split on after 1min. Normal sample purge/desorb produces the problem. I've exchanged purging units with known quality equipment. Same result.
I've always used a narrow bore split liner and a split ratio for volatile analysis. So I'm treading into new territory here =)
Any thoughts on cancellation of error with the septa shavings in the splitless liner?