I am working on a CP3800 varian GC with a saturn 2000 MS.
Does anybody know how I would do manual injection on this system. It uses an autosampler, but due to small sample volumes, I would like to manually inject my samples.
Thanks Aman

In the method configuration set autosampler to none. You can then run a sequence in the normal fashion but the instrument will only start when you press start or, more likely, when you do your injection and start the run via the micro-switch on the inlet.

Rich
"Can't be king of the world
if you're slave to the grind"

It depends which autosampler you have... If it's the 8400 or 8410, in your method, under sample delivery, set the autosampler status to none. If you have an 8200 autosampler, in your 2000 MS control window, go up to the configuration tab and pull the 8200 icon down to the available module window (brings it offline). When you want to return to automation, pull the 8200 icon back up as an available module and be sure to put it in the autostart box. Good luck!
3 posts Page 1 of 1

Who is online

In total there are 182 users online :: 1 registered, 0 hidden and 181 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Google Adsense [Bot] and 181 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry