You people are great - 3 replies in one day.
The new detectors did pass the Varian specifications. The Varian guy was completely puzzled about why our old one was so much more sensitive, and could offer no explanation. He wasn't a typical service engineer / electronics guy, but a specialist troubleshooter. I was quite impressed by his knowledge, and I've been doing GC for a couple of decades. He really knew his product.
The new ones were labelled as having the same milli-Curies as the old one.
You're right, it is a hassle shipping them internationally, but I wanted to see if there was anywhere that could help. So far, I have one name, of a place in North Carolina, who I haven't been able to contact yet. I only have a phone number, and the time difference makes it difficult.
If you know of any others that offer re-foiling or cleaning, please pass on the name or contact details. E-mail address would be really good, from this time zone.
We're fairly sure it was contaminated from the Helium cylinder we were using at that time; done all kinds of troubleshooting. It had really weird symptoms at first, so we didn't figure it out for more than a month; plenty of time for lots of contamination. This is a routine QC lab, with 20 - 40 samples per day, for FID & ECD components, so we have to keep going every day, and get at least some results out.
Unfortunately, the contamination got all through the GC lines, electronic flow controller, filters etc, and it took ages to replace and/or clean everything. We still see the occasional lump on the FID side, but we haven't cleaned anything there, as it is just a minor thing compared to the ECD.
Thanks for all your helpful suggestions. I look forward to the contact details.