Posted: Mon Apr 19, 2021 10:27 pm
Well, after I succeeded in RT Locking my 5973 mass spec for my 8260 method; I went ahead and set it up for SIM/Scan. I got the 69 compounds including the ISS, in 21 groups. Have a fairly good cal and put vinyl chloride and the gases, DBE, and DBCP on SIM-ion and the rest of the VOC compounds on SIM-scan. Using SIM-scan gives about 35% more area counts on each peak compared to simple scan mode. I believe I specified a minimum of 12 points per peak. I had to manually split in half the cluster of compounds preceding and including Flurobenzene.
I have not done it yet; but I am pretty sure I can process the calibration files with my BTEX method and make a cal curve. However, because the BTEX method only has 17 analytes (and the same ISS). I would not be able to run the BTEX method and process it with the 8260 method. It seems to me I will have to run everything in 8260 mode and then process out BTEX if that is all the customer wants reported. That will be a little hard on the filaments because the 8260 method starts at 0.8 minutes (versus 1.6 minutes for BTEX), but no difference otherwise.
I have not done it yet; but I am pretty sure I can process the calibration files with my BTEX method and make a cal curve. However, because the BTEX method only has 17 analytes (and the same ISS). I would not be able to run the BTEX method and process it with the 8260 method. It seems to me I will have to run everything in 8260 mode and then process out BTEX if that is all the customer wants reported. That will be a little hard on the filaments because the 8260 method starts at 0.8 minutes (versus 1.6 minutes for BTEX), but no difference otherwise.