Page 1 of 1

Autosampler error

Posted: Tue Aug 04, 2009 4:15 pm
by chemwipe
I keep getting this error (5890 GC, 7673 injector):

HP3365
Failed to put vial down at front injector
Retry Cancel

Of course I click on "Retry" and the sequence continues on - but this has happened about 3 times this morning already.

There was a large storm here last night which affected some of the network (data system, email), so maybe a power surge/burp?

Any suggestions on what I should check to prevent this from happening?

Thanks,
John

Posted: Wed Aug 05, 2009 2:21 am
by Consumer Products Guy
Shut everything down, including GC and control box. Shut down computer. Start all. We just had a similar issue. I even had last week the 7673 drop a vial, had to look for it. It sure couldn't then put the vial down !!!

Posted: Wed Aug 05, 2009 12:29 pm
by chemwipe
I went ahead and shut everything down yesterday (restarted computer, GC and control box) and started another batch of samples.

Came in this morning and I got the same error after the first sample ran(out of 30!), which apparently happened about 15 minutes after I left the building yesterday, d'oh!

:x

John

Posted: Wed Aug 05, 2009 1:48 pm
by Peter Apps
Hi John

Is this a real error or a false one ? in other words is the arm really failing to release the vial, or is it releasing the vial but its little silicon brain thinks that it isn't ?

In the first case I'd check for mechanical faults. Given the probable power spike the second one might need a software reload, or some new electronics :cry:

Peter

Posted: Wed Aug 05, 2009 5:25 pm
by chemwipe
Is this a real error or a false one ? in other words is the arm really failing to release the vial, or is it releasing the vial but its little silicon brain thinks that it isn't ?
I actually never saw what happened when the error occurred.

I restarted everything again, and when I started up the instrument, the software dialog box read "Injector not ready."

The cable at the back of the controller was loose. Instrument is running fine now.

:roll:

With past experiences, I'd rather this be the problem rather reloading the software (do you know where yours is? haha) or installing new circuit boards

John!