krickos » Wed Aug 05, 2009 4:23 pm
They are impurities alright – no doubt about it.
Incidentally you might like to try other optimization parameters than flow-rate, which seldom has dramatic influence on the separation (within reasonable limits).
Try to play around with pH, column temperature and organic modifier!
Hi
first an apology to grzesiek, due to my perhaps rash wording.
Otherwise, are we rushing ahead here or?
Omar, if you overlay the test solution and SST chromatograms with the Waters column and the test solution do not completly overlap the main peak in the test solution where impurity C is not present, can one really disregard the column factor?
In other words if the "standard colum"/method shows no signs of impurity C in the test solution and the SST shows enough resolution, is additional impurities really the first "thought" or is "column issues" more likely when using zorbax column?
Might add that the SST solution is typically more diluted than the test solution, also naming of peaks seems weird in the zorbax chromatogram (Mesna a smaller riding peak?).
Can you inject Mesna from another source (ie BP, Ph Eur not your source) at the test solution concentration and see if the same/similar thing happens?
I am not ruling out other impurities in Omars "source"/sample but an open mind might be needed.