Need Help on Column life and Peak Splitting
Posted: Mon Aug 03, 2009 4:11 am
Dear All,
I developed a HPLC condition by using the neutral buffer. The mobile phases are:
A: 50 mM phosphate, pH 7.2 (not filtered),
B: 10% tert-butanol in dioxane
Gradient starts at A/B = 90/10, ends at A/B = 35/65 in 55 mins. Then go back to starting condition and equilibrate 10 min.
Temperature at 45 degree
Samples are very clean (USP standards dissolved in water/acetonitrile = 1/1)
Column used: MAC-MOD ACE 3 C18, 150X4.6 mm.
The separation was great at the first 50-60 injections. However, after that I can see all peak heights decrease gradually as injection number increases. Around 80 injections, I can see peak splitting. I have seen this phenomenon on several new columns. However, retention time of all peaks ar unchanged and system pressure only increased about 80 psi (from 1950 to 2030)
Can anybody tell what might be the problem? I guess it is due to the clogging of the inlet frit of the column. I have try to back flush the column with 50/50/0.1 water/acetonitrile/acetic acid. But I cannot get the performance back. I open the nut on the column head, the inlet frit is embedded in a ferrule (seems like one piece). Can anybody tell me where to purchase this kind of unit to replace the old one.
If you have other ideas about what might be the cause, please reply my post
Thank you very much!