Natural ligand lost after HPLC
Posted: Fri Jul 31, 2009 8:40 pm
I'm trying to purify a natural peptide ligand from brain tissue. This can be challenging because I am starting with, at most, 10 pmol. Someone (Uwe?) used the analogy of finding the needle in the hay stack. That is exactly what it is like.
After 5 or so HPLC steps, the purification was going well; much of the protein/peptide contaminants removed. However, when I got down to the 2.1 mm ID sized cyano column, I lost my sample following HPLC (I could no longer detect it). I have a sensitive assay that requires me to burn < 1% for analysis, but I analyzed up to 10% of each fraction and still saw no activity (i.e., it is gone.. or, the machine "ate it.")
I was able to use the same cyano column with success a few days before to purify a different (but similar) natural ligand from the same brain tissue (actually, both ligands were put through the exact sequence of steps). However, when I took the active fractions from this step, diluted with 2-3 volumes water, and injected them on a 1 mm C8 column. Once again, I lost my activity, but on a different column.
This makes two columns that I have lost activity on. Yet, I have used both of these successfully in the past to purify natural ligands, so I'm not convinced it is the column. In fact, this is the first time EVER in ANY HPLC step that I have lost my activity.
There are two differences between this purification and ones in the past:
*I started storing my fractions with 0.% HFIP (to prevent adhesion to the PP tube during storage). (I started this at the beginning, when everything was working fine). I have always kept the fractions at 4C until the next step. I never dry anything down before re-injecting; I just dilute with 2-3 volumes of water.
*I am using a different sample loop. I had been using a 2 mL stainless steel loop. However, in order to inject larger volumes, I borrowed a used 5 mL stainless-steal Rheodyne loop from a neighboring machine.
All else, including instrument, columns, mobile phase, solvent program, etc, is the same. (I -did- change the pre-column frit at the beginning of the purification. )
Could it be something with this loop? (I DO make sure the loop is emptied before re-injecting sample/starting run). I'm thinking, perhaps, it is corroded, and may trap the ligand now that my peptide is relatively pure (and ultra low-abundance). Perhaps this was not of consequence in earlier steps because the impurities were of high enough abundance to prevent such occurrence.
One suggestion I got was to passivate the system with 6 M HNO3 and to flush the columns with EDTA over-night to remove traces of metal ions. I will do this over the weekend. Does anyone have additional advice?
After 5 or so HPLC steps, the purification was going well; much of the protein/peptide contaminants removed. However, when I got down to the 2.1 mm ID sized cyano column, I lost my sample following HPLC (I could no longer detect it). I have a sensitive assay that requires me to burn < 1% for analysis, but I analyzed up to 10% of each fraction and still saw no activity (i.e., it is gone.. or, the machine "ate it.")
I was able to use the same cyano column with success a few days before to purify a different (but similar) natural ligand from the same brain tissue (actually, both ligands were put through the exact sequence of steps). However, when I took the active fractions from this step, diluted with 2-3 volumes water, and injected them on a 1 mm C8 column. Once again, I lost my activity, but on a different column.
This makes two columns that I have lost activity on. Yet, I have used both of these successfully in the past to purify natural ligands, so I'm not convinced it is the column. In fact, this is the first time EVER in ANY HPLC step that I have lost my activity.
There are two differences between this purification and ones in the past:
*I started storing my fractions with 0.% HFIP (to prevent adhesion to the PP tube during storage). (I started this at the beginning, when everything was working fine). I have always kept the fractions at 4C until the next step. I never dry anything down before re-injecting; I just dilute with 2-3 volumes of water.
*I am using a different sample loop. I had been using a 2 mL stainless steel loop. However, in order to inject larger volumes, I borrowed a used 5 mL stainless-steal Rheodyne loop from a neighboring machine.
All else, including instrument, columns, mobile phase, solvent program, etc, is the same. (I -did- change the pre-column frit at the beginning of the purification. )
Could it be something with this loop? (I DO make sure the loop is emptied before re-injecting sample/starting run). I'm thinking, perhaps, it is corroded, and may trap the ligand now that my peptide is relatively pure (and ultra low-abundance). Perhaps this was not of consequence in earlier steps because the impurities were of high enough abundance to prevent such occurrence.
One suggestion I got was to passivate the system with 6 M HNO3 and to flush the columns with EDTA over-night to remove traces of metal ions. I will do this over the weekend. Does anyone have additional advice?