Kwet » Thu Jul 30, 2009 11:19 pm
Which PAH are you analysing?
Usually 16 PAH's are analysed but with your elution beginning with 100 % Ch3CN is not possible! The first 8 components (till pyrene) will elute in void volume (or nearly). Heaviest components (from benzo(a)pyrene to indeno(1,2,3)cd pyrene) can be separated with your isocratic program.
A difference in rentention time from 10.7 ' to 5.6' cannot (only) be due to a temperature difference (Personnaly, I saw retention time shift less than 1 minute between 30 and 40°C).
I think it should have been a chromatagraphic problem. I advice you to compare the pressure between your different injections. It should be stable. If not the system is not enought conditonned or perhaeps your sample can clog the column or frits. Also perhaeps your column is simply dead!
It's not possible to elute the 16 PAH's from your column with your conditions in 10 minutes. 21 minutes is more reallistic (a bit high, I would expect 18 min). I don't understand your fluorescent program. You speak about a run in 10 minute and the fluorescent program change at 12, 14 and 31 minutes! Please check your data, it should be a mistake somewhere.