Cleaning Lipid Contamination with Organic Solvents (HPLC)

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
I am currently using a Agilent Technologies 1200 Series HPLC with an RPC column, which was previously used for analyzing organic compounds containing high concentrations of lipids. I am getting a lot of noise in my analysis even after the standard wash procedures. The solvents used in my HPLC are (Solution A) 5%ACN/95%H20 and (Solution B) 95%ACN/5%H20.

I am currently replacing Solution B with 95% dichloromethane but this does not seem effective so far.

Any recommendations?

If you replace the column with capillary tubing: is the noise better, worse, or the same? That should tell you a lot.

A bit more detail would be useful:

is your method isocratic or gradient?
what detector? if UV, what wavelength?
is the noise "random", or is there a pattern to it?
do you get the same noise pattern when you inject a blank?
if random, is the noise higher or lower under no-flow conditions?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
3 posts Page 1 of 1

Who is online

In total there are 29 users online :: 1 registered, 0 hidden and 28 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot] and 28 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry