Perhaps, I am asking for something obvious, but the following question is not clear to me.
We have a system of HPLC + MS triple quad with ESI source. When we develop a MRM method for a new compound, at first we perform MS1 to determine M+H ion and then we perform MS/MS to get a fragment spectra for this parent ion. For both these experiments we use 10000-100000 ng/ml solution to get enough sensitivity.
After that we make an MRM method using the determined parent mass ([M+H]) and, for example, 4 daughter masses (F1, F2, F3, F4). And we inject solutions in concentration range 5-1000 ng/ml to see that this method works well and to find out the LOD and etc.
The question is:
Why the sensitivities we have for these 4 pairs [M+H]/F in MRM mode do not always correspond to the intensities of respective peaks in MS/MS?
I mean that why, for example, we can have in MS/MS
I(F1)>I(F2)>I(F3)>I(F4)
and in MRM
S([M+H]/F2)<S([M+H]/F1)<S([M+H]/F4)<S([M+H]/F3)
??? (S is an area of a respective peak on MRM-chromatogram)
Dwell times are equal for all MRM pairs.
Thank you for your comments in advance.
Best regards,
Elena