I can't open the images here at work, but if these are making defined peaks then it means the contamination is prior to the column, or the column itself. If you use another column do you get the same contaminates showing up?The title is ambitious, probably. I have a situation: In January 2021 I was transferred to this equipment, because former analyst left a job that month. The problem is that in the past year this machine was heavily used in pesticides analysis, and now it is just impossible to get rid of several pesticides: foramsulfuron, fenuron, bifenazate, nicosulfuron, imazalil, and several others. I tried almost anything to clean the system. I have cleaned capillary and it goes well with autotune. Cleaned column with any variant described in literature. Cleaned system with warm water (70 - 80 C) without column and flow 5 mL/min. Many other things, but those guys just stays in chromatogram. It is indicative, and you can see in TIC picture, that my problems are strict between 4 and 5th min, there are just a few latter on. Please help
Thanks @James_Ball for fast reply. Yes, I have borrowed identical UHPLC column (Zorbax C18, but cannot recall exact model from memory here at home), and same problem was there. Initially, I thought that my former colleague overkill column with pesticides. She played at insane concentrations of pesticides standards at ppm levels. So I have had cleaning the column indefinitely with all combinations of solvents compatible with column. Then I thought, OK, it must be the ion source, so I cleaned capillary, but obviously that didn't solve anything.If you use another column do you get the same contaminates showing up?
Thanks! I tried isopropanol, that was the very first attempt, when I thought it was column the problem. But, then again, I'm not sure that I ever tried flush with solvents without column... Hm, I'm now at the lab, but the instrument is heavily occupied with some cannabinoids in cosmetics samples, tomorrow I'll flush my entire shift, and I'm here 12h a day since relocated to this labHave you tried pumping 100% Methanol or Isopropanol through the system overnight?

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.