problem with an OLD chart recorder
Posted: Tue Jul 28, 2009 6:30 pm
Dear All
I hope my question will not confuse people.
We have an OLD chart recorder which use pen. Recently the recorder can not use the full page (i.e., the pen stop moving upward whenever the absorbance is above 0.3, as a result, the peaks all look like a plaetau).
Because the recorder is too old, no one in the lab knows what could be the problem. W don't even know the model name of the recorder.
So I went online and found the following article:
http://caspar.bgsu.edu/~courses/HPLC/HPLCManual.html
The article says:
"ED LAB: The setup of the chart recorders is a story of it's own. The recorder dealing with the EC detectors displays the output of two channels, namely of detector 1 and detector 2. For both channels it is critical that the gain on the chart recorder itself is calibrated. Calibrate by using the calibration buttons for the two channels on the ESA detector module. For each channel, press the respective button and adjust the gain knob on the recorder until the trace is offset by approx. 1/10 of the width of the page. (Now never, ever touch this knob again!). To zero the trace set the channel on the chart recorder to calibration, put the trace where you want it using the horizontal adjustment knob and zero it with the zero button on the detector module, then set the chart recorder back to 10mV."
Suppose our chart recorder is the one described in the article and that you have ever used the recorder before, what do you think our problem is (recorder can not use the full page) ? How to fix the problem?
Thank you very much for your time!
The article says: