Can I use hexane and dichloromethane as mobile phase with a c18 column ? :?:
You can, but that doesn't mean you should (and I don't know why you would want to).

Assuming you are using a reasonably state-of-the art column, you will void any warranty but you probably won't damage the column, but (and it's a big but!) you will have massive equilibration issues. You will have to flush the column extensively with a co-solvent (most commonly this would be isopropanol) to remove all traces of the default mobile phase/storage solvent before switching to your hexane/dichloromethane and allowing that to remove the iPrOH.

The other concern is with the HPLC system itself. It's not uncommon for manufacturers to use different seal materials for aqueous vs. non-aqueous use, so check with your vendor to confirm compatibility.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 391 users online :: 0 registered, 0 hidden and 391 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 391 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry