Anyone who has used a Waters 6000A solvent delivery system with AR grade solvents ( chromatography grades were only available from Waters at exorbitant premiums ) will attest to the value of vacuum degassing or helium sparging. The first two hours of the day could be spent priming the pumps if vacuum degassing wasn't used.
The idea was to premix solvents 90:10 for A, 10:90 for B, and vacuum filter through 0.45um. That would greatly eliminate bubbles caused by the lower solubility of gases in the blends, as there usually are more dissolved gases in the organic component than the water.
I preferred helium sparging, as it had less effect on the composition, and many of the system had aftermarket pressure relief values that kept a head of helium over the mobile phase - although Dionex had helium system inbuilt.
These days, vacuum filtration has just about gone the way of the dodo. I found alot of people broke the vacuum at the source, thus sucking back the debris into the solvent, unles they had a 0.45 um filter on the vacuum flask outlet. I've seen red rubber chunks from vacuum hoses in solvent reservoirs in some labs.
Milli-Q water systems supply on demand, and prefiltered HPLC grade solvents are relatively cheap. As with other users above, I may filter buffers prepared with salts, but even they are of much better quality, and a 47mm Durapore filter can last me months now, rather than days.
I degas solvent mixtures using an ultrasonic bath. I also have short lengths of 010"ID PTFE tubing as the sole vents on my sealed reservoir bottles, which hopefully also greatly decreases diffusion of air gases, thus reducing load on degasser module.
Please keep having fun,
Bruce Hamilton