I've thought about making injections of lots on both methods and compare their area%.
Thanks,
Denise
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
That’s exactly the way to do it. You don’t need to rationalize as to why the values obtained with the “newâ€I've thought about making injections of lots on both methods and compare their area%.
^^ pretty much what I was thinking - this is for impurities and their responses relative to main peak, so LOD/LOQ are going to be important. I agree that revisiting PDA and runnign some samples via both methods would be a good idea.The cutoff could be important in terms off LOQ/LOD. If the detection is at 210 nm or 220 nm and the mobile phase switches from ACN to MeOH or something else I would make sure that the LOQ is still sufficient. That could be done even by diluting sample solutions.
Aditionally I would run the methods with a PDA to see wether the absortioption maxima change or not (assuming that detection is at the maxima).
Alex
If you get the same response ratio between the impurity and the API using each of the two methods, I think you're probably OK...We also used a lot with significant levels of the impurity to test for stability, ruggedness, reproducibility.
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.