waters 2695 module priming problem
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
6 posts
Page 1 of 1
I have been trying to revive a waters 2695 separation module that has not been used for over a year. I could only withdraw very little solvent with a syringe when I first started to DRY PRIME and no liquid coming out of the waste line when I pressed to continue DRY PRIME. Similarly, nothing came out of waste line when WET PRIME. Any suggestions as to where to start to unclog the lines?
-
- Posts: 11
- Joined: Wed Jun 03, 2009 4:16 pm
Try changing the check valves. When mine got clogged I had the same symptoms - I couldn't pull solvent out at all. Once changed, I was able to get it primed easily.
-
- Posts: 25
- Joined: Mon Aug 30, 2004 2:08 pm
Are you getting any solvent up to the gradient prop. valve? If so, disconnect the line into the inlet check valve and verify flow there. You need to break it down into small pieces and verify flow at each point. If you are getting flow to the inlet check valve, are you getting flow out of the head? If not, the check valve sugestion is probably the first thing to do. When they are left sitting, the ball in the CV tends to stick to the seat. sometimes, hitting them sharply with a wrench will loosen them, but it is probably better to change them out or sonicate them to clean as well as possible. Follow the flow path and check flow at each point. You may also take the frits off the ends of the solvent lines to ensure they are not plugged too. Good luck.
-
- Posts: 2
- Joined: Thu Aug 06, 2009 3:10 pm
Lube Chemist & lcguy1:
You are absolutely right. Check valves were bad. I cleaned them out and now this HPLC flows well.
Had another problem though. The system psi fluctuated quite a bit (usually in hundreds) but the baseline of the chromatogram seemed to be good and flat. Does this mean there is still bubbles trapped in lines or is it normal?
You are absolutely right. Check valves were bad. I cleaned them out and now this HPLC flows well.
Had another problem though. The system psi fluctuated quite a bit (usually in hundreds) but the baseline of the chromatogram seemed to be good and flat. Does this mean there is still bubbles trapped in lines or is it normal?
-
- Posts: 426
- Joined: Mon May 19, 2008 10:09 am
could be bubbles, or a leaking seal.
Try wet priming with methanol (7.5ml/min for 3min).
If not solved, you could take a look at the outlet of your sealwash, if this is moving while pumping (and not only when the sealwash is flushed), you probably have a seal problem.
Ace
Try wet priming with methanol (7.5ml/min for 3min).
If not solved, you could take a look at the outlet of your sealwash, if this is moving while pumping (and not only when the sealwash is flushed), you probably have a seal problem.
Ace
-
- Posts: 25
- Joined: Mon Aug 30, 2004 2:08 pm
So, did you clean the CV's or replace them. Sounds like one, or both, of them are still not working corectly. Probably should just get new ones and replace them. Make sure they are oriented correctly when you put them back in.
6 posts
Page 1 of 1
Who is online
In total there are 24 users online :: 1 registered, 0 hidden and 23 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot] and 23 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot] and 23 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science