response

Do you know whether your parent compounds even elutes with your mobile phase conditions? The parent is quite likely more hydrophobic than its oxo metabolite. If you need 90% MeOH to bring the metabolite off a C18 column, maybe you should try a C8?
All standard disclaimers apply. My posts are my opinions only and do not necessarily reflect the policies of my employer.