7890B GC front FID detector does not ignite
Posted: Sat Mar 06, 2021 10:07 am
by Fouadm
Hi
Many thanks to you in advance.
I have a problem with the front FID which do not go on as the igniting trials have not succeeded.
I have a doubt about the front inlet configuration.
Any suggestions are highly appreciated.
Re: 7890B GC front FID detector does not ignite
Posted: Sat Mar 06, 2021 11:53 am
by dblux_
Hi
1) Does your igniter glow ?
2) Check if hydrogen and air go to FID. Don't rely on setting values. You must prove by flowmeter or by other means. This will check if FID nozzle is not clogged as well.
3) Set stoichiometric flows of gases. Reduce makeup and column flow.
4) Blow a little when igniting.
It must ignite.
BTW - let hydrogen flow as long as air is blown from the line.
Posted: Sat Mar 06, 2021 4:23 pm
by Consumer Products Guy
Sometimes, on certain of our GC systems, we would blow gently into the detector as it was trying to ignite.
Posted: Sun Mar 07, 2021 11:22 pm
by dblux_
That was point 4).
Needless to say FID should have at least 150 degC to ignite - this will be point 5).
Posted: Tue Mar 09, 2021 7:47 am
by dblux_
Let me add point 6) - Lit offset for FID shouldn't be set too high.
Posted: Wed Mar 10, 2021 8:35 pm
by AICMM
Very important point. Blow gently ACROSS the top of the detector, never down into it. Every once in a while they spit fire and you do not want youR head above that. If it makes you nervous thinking about it, you can also use an old fashioned red pipette bulb (the big ones for volumetric pipettes) and puff away at the top of the FID to help it light.
Best regards,
aicmm
Posted: Wed Mar 10, 2021 10:43 pm
by James_Ball
Hi
What would be your doubt about the front inlet configuration? Just wondering how that might affect the detector end.
If the glow plug is going bad, you can try to light it the old fashioned way with a welding torch striker or butane lighter held above the outlet, should get the pop sound when it lights.
Also you can hold a beaker just above the outlet and see if condensation forms, if it does then it is lit but the electronics are not reading signal.
Posted: Fri Mar 12, 2021 12:53 pm
by Peter Apps
Is the column connected to the detector ?
What flows do you have set for air and hydrogen, and are you using make-up gas ? Have you verieid the gas flows with an independent flow meter ?
Have you checked the position of the tip of the column relative to the FID jet - the operator's manual will tell you how far past the detector nut to insert the column.
Posted: Sun Mar 14, 2021 2:51 pm
by Podor
Good point, one time our 7980 wasn't staying lit because the offset was 2 pA, but our baseline was below that. Sometimes using high purity gases with appropriate filters results in too high of a purity and the GC thinks the FID isn't lit.
If you monitor the baseline before the detector shuts down and determine this is the cause, the lit offset is controlled on the GC's interface. There is no way to change it in ChemStation.
Posted: Mon Mar 15, 2021 12:51 pm
One could with the older Chemstations; don't know about the newer versions.