chemstation.. adding peaks/averging noise from2 chromatogrms

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
hi..

could anyone tel me please..
if agilent chemstation software allows..
1. addition of two chromatograms?
2. averaging noise from two chromatograms.. two or more..

or is there any software that would allow this..?
thanks..
Explore!

(1) so far as I'm aware, Chemstation only allows subtraction of chromatograms, not addition.

(2) I wouldn't recommend it anyway, as small variations in retention time (leading to not-quite-aligned peaks) can lead to very strange effects at the start and end of a subtracted/added/averaged peak.

(3) you could, if you really really want to try this, export your chromatogram from Chemstation (in file menu), open in Excel, and do the averaging there. If you're working in LC-MS, I'd suggest using a package such as xcms to look at the data exported into .cdf, and do the averaging there, but this is rather specialised and hard work.

(4) but why not just integrated the peaks as normal, and average the peak areas, rather than averaging the raw data?

I suspect you are doing this because the chromatogram is so noisy that integration is difficult. If this is the case, I would generally try the following steps:
(a) can I increase signal:noise ratio, e.g. by using more sample, choosing my wavelength carefully, using a different detector, etc. etc.
(b) if not, then can I use smoothing? Some people treat smoothing as "manipulating the raw data", but since the data are already smoothed electronically (what do you think the detector's time constant or time-filter is doing??), this is an artificial argument. It's certainly safer to use a controlled numerical filter on the data than to average between two chromatograms of unknown comparability.
2 posts Page 1 of 1

Who is online

In total there are 26 users online :: 1 registered, 0 hidden and 25 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 25 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry