non-repeateble signals GC/MS
Posted: Sat Jul 18, 2009 7:09 am
Dear all,
I use Thermo GC/MS system included DSQ MS with single quadrupole. I work in NCI mode. I use column 15 m Agilent DB-5 ms connected to 1.5 m precolumn with the same phase. Injections are in PTV injector in large volume mode with increasing temperature from 50 to 325 degrees. Column is heating from 80 to 320 degrees. I use liners with glass wool. This configuration is used for 2 years without any changing for pesticides determination.
Lately I noted that I have non-repeatable peaks areas for some compounds. It happens that I make 2 injections of standards mixtures and for the second one I observe much lower signals (300-500% difference). It refers to parameters as follows: chlorthalonil, cypermethrin, captan. In addition for some pesticides (e.g. iprodione, fenhexamid) I cannot observe any peaks although previously I had quite good signals. It is interesting that for these anaytes I have very good signals for prepared samples whilst for standards they are poor. I can observe very good resolution for the rest of pesticides without any drifting of retention times.
MS seems not to be origin of my problem. All the 3 modes (EI, NCI, CI) give good spectra during the tuning without any discrimination. I suspected problems with column but after changing for the new one the problem is still the same. Cleaning the injector with isopropanol did not help too.
Have you any ideas what could be the reason of such situation and how to ressolve this problem?
I use Thermo GC/MS system included DSQ MS with single quadrupole. I work in NCI mode. I use column 15 m Agilent DB-5 ms connected to 1.5 m precolumn with the same phase. Injections are in PTV injector in large volume mode with increasing temperature from 50 to 325 degrees. Column is heating from 80 to 320 degrees. I use liners with glass wool. This configuration is used for 2 years without any changing for pesticides determination.
Lately I noted that I have non-repeatable peaks areas for some compounds. It happens that I make 2 injections of standards mixtures and for the second one I observe much lower signals (300-500% difference). It refers to parameters as follows: chlorthalonil, cypermethrin, captan. In addition for some pesticides (e.g. iprodione, fenhexamid) I cannot observe any peaks although previously I had quite good signals. It is interesting that for these anaytes I have very good signals for prepared samples whilst for standards they are poor. I can observe very good resolution for the rest of pesticides without any drifting of retention times.
MS seems not to be origin of my problem. All the 3 modes (EI, NCI, CI) give good spectra during the tuning without any discrimination. I suspected problems with column but after changing for the new one the problem is still the same. Cleaning the injector with isopropanol did not help too.
Have you any ideas what could be the reason of such situation and how to ressolve this problem?