I know that LC/MS if the ion source is ESI, optimal flow rates are in the 200-300 µL/min range. But this also means that my gradient is delayed by 3-5 minutes plus the column's own void volume - so more like 4-6 minutes. Even if I set up a gradient to run starting at t=0.01, the sample will still experience 4-6 minutes of the initial isocratic phase. And after the gradient is over I have to wait several more minutes to let the back pressure reach an equilibrium state.
We do have access to 2 other isocratic pumps, however. And from what I can tell, an optimal method would involve eluting a short low-organic segment to waste (to wash out the matrix components) followed by a higher-organic segment to elute the analytes into the MS (monitoring by MS/MS, paramters already known).
So I was thinking what I might do is use the switching valve of our MS to program (for example) 2 minutes at 15% methanol and then step up to 5 minutes at 60% methanol, no gradient. This would have much better throughput than programming it all on the single gradient pump.
Would the sudden solvent change have a negative effect on the column? Would it help to provide a small mixing volume (not as large as the problematic 1 mL of course)?
What if I'm just using a small guard column - is that going to be more sensitive to the sudden change?