hello,

I just want to consult regarding my validation method in multiresidue analysis of pesticides in GCMSMS. I computed my LOD and LOQ based from SD in linear regression. I got 4 ppb for LOD and 12 for LOQ. My problem now is that thesignal to noise ratio is below 10 because of the matrix peak beside the peak of the analyte. The recovery and the RSD passed the criteria of 70-120% and RSD of below 20. should I increase it my LOD/LOQ until I get a SN of 10? is matrix peak considered as baseline noise? Thanks
If you're looking for LODs/LOQs in MS data then you can forget signal/noise ratios as there can be zero noise in MS, and work only from the SD. It is more referred to as MDQ (minimum detectable quantity) or similar.
Where can I buy the kit they use in CSI?
An obvious peak would not be considered baseline noise in my opinion. If you can not separate the analyte from the matrix peak completely then that would be a reason to question the detection limits, but as long as they can be separated I would just look at the noise in the baseline versus the height of the actual peak.

As said above, for MS it is often difficult to do S/N when noise can be near zero, I normally set a peak area limit of something like 1000 area under the peak( or whatever gives a good peak in your instrument) and back calculate to what concentration should give me that area, then inject that concentration and see if it works.
The past is there to guide us into the future, not to dwell in.
The most rigorous and robust estimate of LOD and LOQ for a method are the concentrations (or quantities) at which the rsd on repeated measurements is 30 and 10% respectively. This is proof positive of how repeatable the method is at that level. The SD of the calibration can be influenced by very good (or bad) repeatability at levels way above the lower limits, and the signal:noise on a chromatogram gives the most optimistic possible estimate of LOD and LOQ, because it completely ignores all the other sources of variation.

Peter
Peter Apps
For a brief discussion of why S/N is not a good judge of limit of detection and quantitation check out http://www.absciex.com/Documents/Downlo ... 059150.pdf
They cover the affect of filtering and smoothing on S/N calculation.
5 posts Page 1 of 1

Who is online

In total there are 30 users online :: 2 registered, 0 hidden and 28 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot], Semrush [Bot] and 28 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry