After maintaining the MSD it's ok, but after a few weeks it comes again. Mostly it can be improved by fastening the column nut of the MSD interface inside the oven. Changing the inlet septa gave no change.
So I assume the problem comes from an air leak at the interface. I know, that the nut has to be retightened several times after installing a new ferrule, but how could I improve the connection more permanently? We use Agilent ferrules made of 85/15% Vespel/Graphit (0.5mm ID, long, 5062-3506), which are recommended for GC/MS according Agilent catalogue. Has anybody a better recommandation?