I would like your advice about the following problem.
I am validating my LC-MS/MS method for a basic compound and its metabolite.
I use a C18 column with a gradient of A) water 25 mM ammoniumformate pH 3 B) MeOH containing 0.1% formic acid.
The mass spectrometer is operating in positive ESI, MRM-mode with all parameters nicely optimized.
Injection of a standard solution gives nice, repeatable peaks.
In the future this method will be used for the analysis of urine and whole blood, so sample preparation is obviously needed. I selected an SPE method using mixed mode SPE cartridges (Bond Elut Certify, C18 and a cation exchanger) and I use the protocol that is given by Varian itself. The protocol also ressembles other protocols used in the literature, so I suppose this hould work fine.
However, when validating matrix effects (ME) and recovery (RE), following problem was seen.
I used a standard, a pre-SPE-spiked sample and a post-SPE-spiked sample to calculate RE and ME.
When using urine, RE and ME were bad (about 30-40%). However when using the ratio analyte over internal standard for calculating ME and RE, values were excellent (90-110%). Moreover, when I repeated the same processes with the same matrix, variability for the pre-SPE-spiked samples and a post-SPE-spiked samples was awful... So the internal standard compensates great for ME and RE, but of course can not solve the impact of the variability on sensitivity.
I repeated this experiment with pure water, in orde to exclude any effect of the urine, but the same was seen: good variability for standard, bad variability for different pre-SPE-spiked samples and post-SPE-spiked samples and significant loss of signal in the pre-SPE-spiked samples AND post-SPE-spiked samples!
I already checked some elements.
The glassware and pipets etc used all work fine.
I tested another SPE-protocol with a different elution solvent, this resulted also in bad variability and loss of signal.
When I simply evaporate some elution solvent and then add the standard, variability is again good and there is some signal loss (5-10%), so the elution solvent doesn't seem to be the cause...
When I simply ran some elution solvent over the SPE-column (so no condition, sample application or wash performed), I evaporate the elution solvent and the add the standard, variability is bad and so is signal loss.
Another person of the lab performed the SPE-procedure, the same results were obtained.
Mobile phase was made fresh, another column was used, ... : I think the LC-MS/MS method works fine...
I really would appreciate some advice since the problem is starting to drive me crazy!
Thanks!