Any tricks to reduce needle plug?

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
Dear all,

I am using manual injection needle (Hamilton, point style 2, needle guage 26 or 22) to inject samples.

Almost everytime I insert the needle through the septum, the needle will "tear off" a tiny bit of septum, causing the plugging of the needle.

Is there any tricks to avoid this problem? (Am I over-tighten the septum retainer nut to cause the above problem?)

Thanks in advance.

Hi Anthony

The point style that you have is the worst option for septum damage. I cannot understand why syringe manufacturers continue to offer them as their default option.

Go for a domed or cone tip, or even a square end with the corners taken off with very fine abrasive paper.

You might also be tightening the nut too hard - make it just tight enough to re-seal after an injection when it has warmed up to working temperature. If you tighten with the septum cold it will be too tight once it has expanded with the heat.

Peter
Peter Apps
2 posts Page 1 of 1

Who is online

In total there are 90 users online :: 1 registered, 0 hidden and 89 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Bing [Bot] and 89 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry