alicee! » Sat Jul 11, 2009 1:09 pm
The potential pitfall I see is this:
In assays. Here you tend to use the "same way" of evaluation as in the intended final method when you do linearity and accuracy.
In impurities you do a linearity investigation of minimum 5 points (think external calibration curve with minimum 5 points) but when you do accuracy, you may use internal normalisation (area %), weight/weight %...
You will likely not use a 5 point calibration curve when you run the method routinely or?.
Hence the result evalution does not match between validation and actual method.
Sure you can study linearty for impurities by spiking impurities at different levels into solutions derived from a stock sample solution and then evaluate linearity and accuracy at the same time but thats not what you asked for.
Or I am missing something?
i have a few doubts very close to wat ur talking abt..
i dont have purified impurities to spike.. so i dunno how to prove accuracy using this approach.
how i plan to do it is using my purified protein.. assuming say an oxidized impurity will not behave very differnt from the native purified form that i m using to establish accuracy..
i plan to do a linearity and backcaluculated graph based recovery..
i want to know.. u mentioned 5 pionts.. wat is the range that you would evaluate this for.. near the impurity range.. i.e if my injection amount was 15 mcg.. do i do it near 0.2-2% of this.. where my impurities would lie.. or 0.2-120% of 15 mcg.. i.e. full range..
is a full range linearity expected?
also..
for quantitation of impurities.. i could do a 5 point curve between 0.2-2% or do an area% like u said.. i get a difference in these values.. as i donot expect full range linearity.. is this unusuall to happen.. what is the preferred approach to address this..
thanks!