vliu » Tue Jun 30, 2009 1:27 am
bisonium:
1. What compound is it you inject (chemistry)?
A: Surfactant, trying to measure both micelle and monomer form
2. What column are you using (packing?)
A: Size Exclusion column 2000
3. Are you sure the column doesnt work as ion exhange and therefor your compound sticks inside the column?[/quote]
A: Not sure.
bhuvfe:
When you blow up the chromatogram do you see noise or is it really a flat line?
A: Flat line, no spike at all.
Vlad Orlovsky:
Lamp reference energy is 200+ and has only used 1000 hours only. I do get a solvent front response when injecting water without column. Will try injecting sodium chloride.
I am using compressed air rather than Nitrogen, I've compared my results using both gases before with a validated method, and the results were comparable.
What can I do if column bleed is causing the flate basline on the ELSD? Does it mean I need a new column?

I am using a guard column already...
I am using 50% acetonitrile with 0.01M ammonium acetate (I've tried other % of acetonitrile as well). I think I should mention that I did get some response using the Refractive detector but it is not sensitive enough for my purpose.
All lines are in MeOh/water.
Koen Hollebekkers:
How did you overcome column bleed?
So you are suggesting testing the baseline with the pump off ?