Why no noise in LC/MS application?
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
11 posts
Page 1 of 1
I have developed a method for a carbohydrate using ESI LC/MS. The compound do not contain any functional groups that easily (de-)protonate so I have used the sodium adduct for its detection. Nor does the compound fragment into ions that can be detected, so I have used SIR for quantitation instead of MRM. My problem is the following: When running sea water samples the matrix effects gets high, which not is strange, but in the same time the noise levels out and cannot be measured. In standard runs the noise looks completely normal. The peaks in the Sea water samples (that has correct retention times compared with standards) have areas that are about 10-1000 times higher compared with the area for a standard peak with an area that equals LOQ when no matrix is present. The method seems to work OK, the problem is that I cannot calculate a S/N ratio in the samples. (I have enlarged the noise with the software, but it still becomes nothing but a straight line). I use a labelled internal standard. The compound is chlorinated, so I aquire the two first peaks in the spectra, all in all three SIR transitions are monited in the same run. I guess cross-talk has nothing to do with this?
-
- Posts: 71
- Joined: Thu Jun 11, 2009 9:42 am
I have a question to start with my guess: Have you an idea of the levels of your analyte in real sea water sample? I mean, if you consider the high matrix effect you've mentioned, is there a correlation between the peak you observe and the possible concentration?
If so, the noise absence could be directly related with the matrix effect that causes a strong ion suppression.
If so, the noise absence could be directly related with the matrix effect that causes a strong ion suppression.
Samuele Pedraglio
Developability Dept.
NiKem Research S.r.l.
Italy
Developability Dept.
NiKem Research S.r.l.
Italy
-
- Posts: 6
- Joined: Wed Jul 01, 2009 9:36 am
Thank you for the fast answer! When quantifying the peaks (relative the labeled standard), I get results that well fits the results of others. I would like to ask if you know if this with no noise due to high matrix effects is mentioned anywhere in the literature? It would be interesting to read a journal article discussing this problem. Thanks again.I have a question to start with my guess: Have you an idea of the levels of your analyte in real sea water sample? I mean, if you consider the high matrix effect you've mentioned, is there a correlation between the peak you observe and the possible concentration?
If so, the noise absence could be directly related with the matrix effect that causes a strong ion suppression.
-
- Posts: 71
- Joined: Thu Jun 11, 2009 9:42 am
I think you'd have no results searching for this issue in literature, my guess is strictly related to your high ion-suppression.
Another idea: are you sure you don't have discharge problem into your source? If you have a very salty sample, and with sea water I guess it is, you could generate discharge and then no signal.
Another idea: are you sure you don't have discharge problem into your source? If you have a very salty sample, and with sea water I guess it is, you could generate discharge and then no signal.
Samuele Pedraglio
Developability Dept.
NiKem Research S.r.l.
Italy
Developability Dept.
NiKem Research S.r.l.
Italy
-
- Posts: 6
- Joined: Wed Jul 01, 2009 9:36 am
Thank you very much your help. The samples originates from the Baltics where salinity is much lower than in southern of Europe and they have also been washed in an SPE-step. It must be that the same masses as my analyte exist in the matrix and constantly reaches the detector that is the cause of this problem. It is a bit annoying, since the method seems to work in every other aspect. Thanks again
-
- Posts: 71
- Joined: Thu Jun 11, 2009 9:42 am
in this case you must have noise and, eventually, a low S/N if you have low analyte levels in your samples.It must be that the same masses as my analyte exist in the matrix and constantly reaches the detector
this should give low, or absent, signal in every conditions.the detector that is the cause of this problem
Samuele Pedraglio
Developability Dept.
NiKem Research S.r.l.
Italy
Developability Dept.
NiKem Research S.r.l.
Italy
-
- Posts: 6
- Joined: Wed Jul 01, 2009 9:36 am
It is true that I have some broad "peaks" in my chromatograms. If these are noise, than I must define LOQ as 2 times S/N (if I want to stick with this method). Areas in quantified samples are at least 10 times larger than the area for a peak that is 10 times S/N in a standard run when no matrix is present. What is your opinion? Would you say this low value for LOQ definition is defendable?
-
- Posts: 586
- Joined: Mon Oct 04, 2004 3:00 am
What is the threshold set at? If you set the threshold too high you can easily lose the noise if there is ionization suppression.
-
- Posts: 6
- Joined: Wed Jul 01, 2009 9:36 am
Great! This seems to do the trick! When playing around with the threshold and smoothing settings the peaks in samples are at least 5 times S/N, after a quick try. Thank you both!
-
- Posts: 1890
- Joined: Fri Aug 08, 2008 11:54 am
this is a nice example of why the S/N ratio approach to LOQ carries significant dangers. It can easily become completely meaningless. I much prefer methods based on the standard deviation of low calibration points.
-
- Posts: 6
- Joined: Wed Jul 01, 2009 9:36 am
I do agree. Still: Quantifying low calibration points when no noise is visible is also a bit scary.this is a nice example of why the S/N ratio approach to LOQ carries significant dangers. It can easily become completely meaningless. I much prefer methods based on the standard deviation of low calibration points.
11 posts
Page 1 of 1
Who is online
In total there are 26 users online :: 0 registered, 0 hidden and 26 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 26 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 26 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science