GC/MS contaminant peak
Posted: Wed Jul 01, 2009 7:26 am
Hi all,
I've got a problem with our Agilent 7890 GC/MS headspace system regarding a persistant contaminant peak which is co-eluting with a peak of interest - but according to the scan the peak of interest (in this case ethanol) is not the contaminant.
It has a consistent RT which led me to believe that it could be an injection problem so i have investigated the following:
- Replace septum/liner: NO EFFECT
- Replace Syringe: NO EFFECT
So i decided to turn my attentions on the column:
-Baked the column as described by Agilent: NO EFFECT
-Trimmed the column: NO EFFECT
-Even installed a new column: NO EFFECT
I am currently checking the vials being used are not causing the problem but i feel that i am scraping the bottle of the barrel now.
Have i missed something?
Any opinions would be welcome.
Thanks
Steve
Thanks
Steve