HPLC/chitosan/pectin
Posted: Mon Jun 29, 2009 3:11 pm
I am having problem with HPLC analysis and I would highly appreciate some advice…
I am performing the dissolution studies of the drug from polymeric matrices obtained by the use of hydrophilic polymers, namely chitosan and amidated pectin. The dissolution media is simulated saliva (phosphate buffer pH 6.5 without enzymes).
The drug quantification was performed by HPLC on Purospher STAR RP-18 column (column size 125/4 mm, granule size 5 microm). The mobile phase is methanol:0.01 M phosphate buffer pH 6.8 in ratio 70:30, flow rate at 1.5 mL/min, detector at 235 nm (absorption maximum of the drug), column temperature 40C, injection volume 20 microL. I have successfully validated the method and I have obtained excellent results.
Than I have started the dissolution studies. And I have noticed that the pressure of the system is constantly increasing… During the validation of the method it was stabile, around 16.5 MPa, but when I started to analyze the samples containing the amidated pectin and chitosan, the pressure was risen up to 25 mPa… And this pressure increase is constant, not only during the injection of the sample. Is it possible that pectin is blocking the column? All samples have been filtrated prior to injection.
The solubility of chitosan in the dissolution media is low; therefore I think that he should not be the cause. I have tried everything, the back-flushing of the column with the water (pectine is soluble in water), the washing of the column according to the manufacturers advice (water-methanol-acetonitrile-methanol-water), have purged the tubes but the pressure is still high. Moreover, I have observed the tailing of the pick (at concentrations that are in the middle of the working range. For the same drug concentration, during method validation period, the peak was symmetrical)…
The mobile phase is stabile, during the 2 days after mixing the components I have not notice crystallization of the salts from the buffer. Addition of the simulated saliva to mobile phase does not change its appearance. At the end of each working day, I have washed the column by 30 ml of methanol/water (70:30) mixture followed by 30 mL of methanol/water (30:70) and finally by 30 mL methanol/water (70:30). The producer of the column has stated that 30 mL is the washing volume for this column… I think that by this procedure I should have removed buffer and remaining polymer by this washing.
Is it possible that pectin is messing up with the silica? Did anybody have similar problems? Any advices about performing the HPLC analysis in the presence of hydrophilic polymers? How, I have ordered the new column, but I have fears not to destroy it… What am I doing wrong?