HELP GCMS Sensitivity Problems Shimadzu

Posted: Mon Jun 29, 2009 12:37 pm
by LWC
Hello,

I have a Shimadzu QP 5000 and recently we had an air leak that lead to several issues. I have cleaned the MS and replaced all of the parts of my injector and trimmed the column a bit on the injector side. My tunes look good and my pressure is steady however I cannot get my sensitivity back to where it once was. Is this a column issue? What do you think?

Thanks

Posted: Mon Jun 29, 2009 12:47 pm
by Consumer Products Guy
Try with a manual injection, different syringe.

Posted: Mon Jun 29, 2009 12:51 pm
by LWC
[quote="Consumer Products Guy"]Try with a manual injection, different syringe.[/quote]

I have, no difference, sorry forgot to include that.

Posted: Mon Jun 29, 2009 3:45 pm
by Ron
What is the condition of the filaments? Have you tried switching to the second filament to see if the response is better? The QP-5000 has a ribbon filament instead of a wire, and if the filament warps or wrinkles the sensitivity can be affected.

Posted: Mon Jun 29, 2009 3:50 pm
by LWC
[quote="Ron"]What is the condition of the filaments? Have you tried switching to the second filament to see if the response is better? The QP-5000 has a ribbon filament instead of a wire, and if the filament warps or wrinkles the sensitivity can be affected.[/quote]


The filaments were new in MArch and switching them doesn't seem to make a difference. Talking to another guy from Shimadzu he lthinks it could be a ferrule issue in the inlet or and electrical phenomenon?

Posted: Thu Jul 02, 2009 5:19 pm
by AICMM
LWC,

Being that one of the last things you did was trim the column, are you sure you got the column position in the inlet correct? Can make a huge difference in sensitivity. Split or splitless and what kind of liner?

Best regards.