What is the typical of %RSD of three triplicate injections from the same sample vial? I am using a Thermo TSQ Quantum LC/MS instrument. I tried to make a standard curve using internal standard. However, the peak area ratio of three replicate (from the same vial, just injected three times successively) was not consistent, the %RSD was 3% to 5%. What is the typical %RSD in similar cases? What are the possible reasons for this? Thanks a lot.

It's quite dependent from your injection and pumping system.
We work with a Waters UPLC and we'd obtain reproducibility around 0.5-1%.
Samuele Pedraglio
Developability Dept.
NiKem Research S.r.l.
Italy

thanks for your reply. Do you think the reproducibility has anything to do with the quality of tuning?

I would say that the HPLC should perform in the <1-2% RSD range of injection reproducibilty. I would calculate this from an n=6 injections though. You should probably take into consideration you're using an LC-MS, so I would not expect to get the exact same reproducibility. This is based on a neat sample too. I don't know if your samples are extracted from something or analyzed neat. If your samples are extracted, I would not be concerned with 3-5% RSD. I can't tell if your issues are from tuning since I don't know enough information on your method.

Do you notice a trend with your sample injections? Does your peak area continue to increase, or decrease? How about retention time? Is the ratio of internal standard area to analyte area proportional?

for HPLC it is 2 or less than 2 % but in case of LCMS/MS it depends on your QA department I mean the STP it can vary from 5 to even 15 %.
5 posts Page 1 of 1

Who is online

In total there are 17 users online :: 1 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google Adsense [Bot] and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry