In-situ isomerisation: calibration curve and LOD/LOQ

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

6 posts Page 1 of 1
I have a compound that is isomerising in solution and the cis and trans isomers can be clearly separated by HPLC. However the trans isomer is about 98% while the cis isomer is about 2%. At low concentrations I am able to integrate the trans isomer easily but not the cis. From literature the equilibrium between the two isomers always stays the same providing the solvent system does not change. Is it acceptable for me to use only the larger trans isomer peak when generating a calibration curve and when determining my LOQ and LOD? If I can't my LOQ/LOD values are going to be very high!

If you are quantitating the total amount of the compound, and if, indeed, you can demonstrate that the equilibrium is 98/2, then yes, you should be able to use just the major isomer for LOD & LOQ estimation. LOQ (S/N = 10) effectively implies a CV of +/- 5%. LOD (S/N = 3) implies a CV of +/- 20%.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

If I had a method that was unable to detect 2% of my load, I would seriously consider a comprehensive revision of the method.

Best Regards
Learn Innovate and Share

Dancho Dikov

If I had a method that was unable to detect 2% of my load, I would seriously consider a comprehensive revision of the method.

Best Regards
But if you have to determine LOD and LOQ for impurity profiling, and should reach 0.05%, then it's possible that 0.05% from 2% isn't visible while 0.05% from 98% still visible.

Ace


But if you have to determine LOD and LOQ for impurity profiling, and should reach 0.05%, then it's possible that 0.05% from 2% isn't visible while 0.05% from 98% still visible.

Ace
I'm not able to understand your thought to say this.
0.05% of 98% is always less than 2% of 100%.
If you're not able to see the 2% you'll never be able to see 0.05%.
Samuele Pedraglio
Developability Dept.
NiKem Research S.r.l.
Italy


But if you have to determine LOD and LOQ for impurity profiling, and should reach 0.05%, then it's possible that 0.05% from 2% isn't visible while 0.05% from 98% still visible.

Ace
I'm not able to understand your thought to say this.
0.05% of 98% is always less than 2% of 100%.
If you're not able to see the 2% you'll never be able to see 0.05%.
What I meant to say: if you have to look for peaks of 0.05% from your main peak, then it's possible that you can't see the 0.05% peaks from your main peak if the 2% peak is the main peak for your degradants you're searching for. So actually you are searching for peaks of 0.05% * 2% = 0.001% instead of 0.05% * 98%= 0.049%

I hope it's a bit clear now :?
Ace
6 posts Page 1 of 1

Who is online

In total there are 3 users online :: 2 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry